how to reduce CV in ELISA


Guest susan
 Share

Recommended Posts

Hi Susan, and welcome to the forum!

I’m sorry to hear that you get high CV%. Before being able to give you a specific answer, we need to know a few more details:

1. Is it the CV% of inter-assay (when comparing different experiments) or intra-assay (within one plate) you are referring to?

2. Is it the standard curve or the samples (or both) that worry you?

3. How high CV% do you get?

4. Are the CV's for OD values or to back-calculated concentrations? Please note that for very low OD values, such as the buffer background, CV values will always be high.

Regardless of your answers above, here are some general things to consider in order to minimize high CV%:

  • Use calibrated pipettes, and preferably the same pipettes in all experiments.
  • If your samples need to be diluted with a high dilution factor, make one or several pre-dilutions, and do not take too low sample volumes (lower than 5 ul).
  • Adjust all reagents, except Mabtech’s TMB substrate, to room temperature before use and preferably incubate plates at the same temperature.
  • Also, all incubation times need to be the same between experiments.
  • Include several replicates of the samples and at least duplicates of the standard (as recommended).

Please also be advised that for human IFN-g (3420-1H-6) and Latent TGF-b (3550-1H-6) we offer validated ELISA Pro kits (3420-1HP-1 and 3550-1-HP-1, respectively) with pre-coated plates, buffers and substrate that ensure reproducibility and thus minimize risk of high CV.

Link to comment
Share on other sites

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

Loading...
 Share

×
×
  • Create New...